Subject:

Re: Negative Data- ICP-MS

From:

Azam Noori <[log in to unmask]>

Reply-To:

Azam Noori <[log in to unmask]>

Date:

Fri, 1 Jun 2018 20:52:10 +0000

Content-Type:

text/plain

Parts/Attachments:

Parts/Attachments

text/plain (109 lines)

 Steven,
Thank you very much for the great comments. I see I need to learn a lot. 
You are right, the internal standard has declined. It should not happen. I will check the neutralizer and change the tubes.
A quick question about the digestion method. In this case I will have HNO3, H2O2  and HCl in tubes. Should I make the rinse solution with the same concentration of HNO3 and HCl or it does not matter. At this point I have 2% HNO3.
I appreciate your time and help.
Regards,
Azam Noori, Ph.D.

Assistant Professor
Department of Biology
Merrimack College
315 Turnpike Street
North Andover, MA 01845
Mailstop: N8
978-837-5744
    On Friday, June 1, 2018, 4:03:19 PM EDT, Pappas, Richard Steve (CDC/ONDIEH/NCEH) <[log in to unmask]> wrote:  
 
 Azam, one more thing. Rob Franks's trifocals are better than mine for looking at data expressed in 16x6 matrices.
He noticed that your Indium internal standard concentrations(?) went negative after the first couple of samples.
It is possible that the undigested plant material clogged the nebulizer or similar blockage. You should check to see if the nebulizer is indeed partially or completely blocked.
It may be that you do not understand the purpose of an internal standard. An internal standard is usually not quantitated by calibration. Its concentration is not increased with higher concentration standards. It is designated as internal standard in the software. Then as signal fluctuations occur as caused by plasma fluctuations, signal drift, etc.; the change in the constant concentration internal standard corrects the calibration standard-calibrated analyte measurements for these fluctuations.
If I understood the matrix correctly (and I was a math minor and took linear algebra), In concentration should not increase from 5 to 10, 25, 50, 75, 100 units, whatever the units are.

R. Steven Pappas, Ph.D.
Team Lead, Tobacco Inorganics Group
Centers for Disease Control & Prevention
4770 Buford Highway, NE
M.S. F44, Building 110
Atlanta, GA 30341-3717, USA

-----Original Message-----
From: PLASMACHEM-L: Analytical Chem.(ICP's, DCP's, MIP's). <[log in to unmask]> On Behalf Of Azam Noori
Sent: Friday, June 1, 2018 2:17 PM
To: [log in to unmask]
Subject: Re: Negative Data- ICP-MS

 Hello All,
Thank you very much for the comments and suggestions. In response to your questions: Bill, I digested plant tissues in 2% HNO3 using hot block. Samples were filtered using Whatman #1. The final volume for all samples were brought to 50ml. I used "In" as an internal standard for Ag with the concentration of 2ppb. Internal standard was added to all samples, blank, and standards. I analyzed Ag 107 as well. The result is similar to Ag 109. The 10ppb standard showed 661ppb for Ag 107. It is not just about Ag, most elements are showing similar data.
Steve, thank you for the article. I agree that it could be related to Ag. I will follow your suggestion to clean the instrument. Laurel, I checked the internal standards. It shows the concentration of internal standards as percentage.  I copy-pasted some of the data here (I am not sure if you can see them). 

|  | Mo 98
 (ug/L) | Ag 107
 (ug/L) | Ag 109
 (ug/L) | Cd 111
 (ug/L) | In 115 (IS) |
| Std1 5 ppb | 5.064 | 5.000 | 5.000 | 5.000 | 98.7% |
| Std2 10 ppb | 9.705 | 9.884 | 9.888 | 9.897 | 99.9% |
| Std3 25 ppb | 25.549 | 25.586 | 25.580 | 25.262 | 90.0% |
| Std4 50 ppb | 49.744 | 49.520 | 49.410 | 49.944 | 92.7% |
| Std5 75 ppb | 75.283 | 77.423 | 77.270 | 75.369 | 89.4% |
| Std6 100 ppb  | 100.216 | 101.912 | 102.669 | 101.297 | 89.1% |
| Ag30-1 S  | 1.257 | 41.048 | 39.526 | 1.134 | 80.9% |
| Ag30-2 S  | 25.396 | 45.313 | 51.452 | 29.933 | 9.3E-5% |
| Ag30-3 S  | 29.705 | 107.315 | 115.015 | 37.079 | 4.8E-5% |
| Ag30-4 S  | 20.048 | 44.295 | 35.463 | 34.051 | 1.0E-4% |
| Ag30-5 S  | 13.070 | 330.094 | 348.927 | 41.944 | 7.5E-5% |
| 10 ppb | 29.326 | 661.546 | 623.595 | 26.292 | 7.9E-5% | Cal Blank |   
| |   |   |   | 100.0% |
| NO10-1 S | -0.544 | -58.368 | -43.304 | 32.019 | 61.7% |
| NO10-2 S |   | -189.509 | -177.342 | 16.160 | 74.4% |
| NO10-3 S | -13.904 | -229.558 | -213.942 | -15.295 | 449.8% |
| NO10-4 S | -14.204 | -220.707 | -207.380 | -17.115 | 531.1% |
| NO10-1 S | -16.075 | -221.138 | -207.707 | -16.105 | 486.6% |
| 5 ppb | -16.204 | -219.315 | -204.345 | -15.270 | 484.8% |


Regards,
Azam Noori, Ph.D.
Assistant Professor
Department of Biology
Merrimack College
315 Turnpike Street
North Andover, MA 01845
Mailstop: N8
978-837-5744
    On Friday, June 1, 2018, 11:18:52 AM EDT, Mc Ellistrem, Laurel J. <[log in to unmask]> wrote:  
 
 Hi-
In case you haven't checked this, can you check the concentrations of the internal standard entered in the file?  Open the file to see the number, rather than trusting the file name.  I had a similar problem that occurred because I had entered the concentrations for the internal standard but apparently hadn't hit "enter" but saved it.  When searching for the problem in the known sample analysis, the numbers for the internal standard were listed as the 1.0000ppb, the default starting value when you open a new standard file.  Good luck, Laurel

Laurel McEllistrem, Ph.D.
715-836-2723
[log in to unmask]
University of Wisconsin-Eau Claire
Materials Science and Engineering Center Eau Claire, WI 54701

-----Original Message-----
From: PLASMACHEM-L: Analytical Chem.(ICP's, DCP's, MIP's). <[log in to unmask]> On Behalf Of Azam Noori
Sent: Friday, June 1, 2018 9:12 AM
To: [log in to unmask]
Subject: Negative Data- ICP-MS

Good morning,

My name is Azam Noori. I am an assistant professor at Merrimack College. We have purchased an ICP-MS lately. It is NEXION 350X (Perkin Elmer). I am working with plants. I analyzed some samples yesterday. The calibration curve for most elements was good (Correlation factor: 0.998-0.999).
The concentration of elements in the first 5 samples was reasonable. Then I used the standard 10ppb as a sample to make sure the instrument is reporting the correct concentration. The concentration of most elements was much higher than 10 ppb in this sample. For example for silver (109), it was 623 ppb while when same sample was used as a standard the concentration was 9.88 ppb. I analyzed some samples after this, and the concentration of most elements (Ag, Mo, Cd, Sn, As, Ni, Fe, Cr, Pb, and V) in all samples were negative. It is not what I expected. Do you have any suggestions for troubleshooting?

Regards,

Azam Noori, Ph.D.
Assistant Professor
Department of Biology
Merrimack College
315 Turnpike Street
North Andover, MA 01845
Mailstop: N8
978-837-5744  
  

Top of Message | Previous Page | Permalink

Advanced Options


Options

Log In

Log In

Get Password

Get Password


Search Archives

Search Archives


Subscribe or Unsubscribe

Subscribe or Unsubscribe


Archives

September 2018
August 2018
July 2018
June 2018
May 2018
April 2018
March 2018
February 2018
January 2018
December 2017
November 2017
October 2017
September 2017
August 2017
July 2017
June 2017
May 2017
April 2017
March 2017
February 2017
January 2017
December 2016
November 2016
October 2016
September 2016
August 2016
July 2016
June 2016
May 2016
April 2016
March 2016
February 2016
January 2016
December 2015
November 2015
October 2015
September 2015
August 2015
July 2015
June 2015
May 2015
April 2015
March 2015
February 2015
January 2015
December 2014
November 2014
October 2014
September 2014
August 2014
July 2014
June 2014
May 2014
April 2014
March 2014
February 2014
January 2014
December 2013
November 2013
October 2013
September 2013
August 2013
July 2013
June 2013
May 2013
April 2013
March 2013
February 2013
January 2013
December 2012
November 2012
October 2012
September 2012
August 2012
July 2012
June 2012
May 2012
April 2012
March 2012
February 2012
January 2012
December 2011
November 2011
October 2011
September 2011
August 2011
July 2011
June 2011
May 2011
April 2011
March 2011
February 2011
January 2011
December 2010
November 2010
October 2010
September 2010
August 2010
July 2010
June 2010
May 2010
April 2010
March 2010
February 2010
January 2010
December 2009
November 2009
October 2009
September 2009
August 2009
July 2009
June 2009
May 2009
April 2009
March 2009
February 2009
January 2009
December 2008
November 2008
October 2008
September 2008
August 2008
July 2008
June 2008
May 2008
April 2008
March 2008
February 2008
January 2008
December 2007
November 2007
October 2007
September 2007
August 2007
July 2007
June 2007
May 2007
April 2007
March 2007
February 2007
January 2007
December 2006
November 2006
October 2006
September 2006
August 2006
July 2006
June 2006
May 2006
April 2006
March 2006
February 2006
January 2006
December 2005
November 2005
October 2005
September 2005
August 2005
July 2005
June 2005
May 2005
April 2005
March 2005
February 2005
January 2005
December 2004
November 2004
October 2004
September 2004
August 2004
July 2004
June 2004
May 2004
April 2004
March 2004
February 2004
January 2004
December 2003
November 2003
October 2003
September 2003
August 2003
July 2003
June 2003
May 2003
April 2003
March 2003
February 2003
January 2003
December 2002
November 2002
October 2002
September 2002
August 2002
July 2002
June 2002
May 2002
April 2002
March 2002
February 2002
January 2002
December 2001
November 2001
October 2001
September 2001
August 2001
July 2001
June 2001
May 2001
April 2001
March 2001
February 2001
January 2001
December 2000
November 2000
October 2000
September 2000
August 2000
July 2000
June 2000
May 2000
April 2000
March 2000
February 2000
January 2000
December 1999
November 1999
October 1999
September 1999
August 1999
July 1999
June 1999
May 1999
April 1999
March 1999
February 1999
January 1999
December 1998
November 1998
October 1998
September 1998
August 1998
July 1998
June 1998
May 1998
April 1998
March 1998
February 1998
January 1998
December 1997
November 1997
October 1997
September 1997
August 1997
July 1997
June 1997
May 1997
April 1997
March 1997
February 1997
January 1997
December 1996
November 1996
October 1996
September 1996
August 1996
July 1996
June 1996
May 1996
April 1996
March 1996
February 1996
January 1996
December 1995

ATOM RSS1 RSS2



LISTSERV.SYR.EDU

Secured by F-Secure Anti-Virus CataList Email List Search Powered by the LISTSERV Email List Manager